Improve / optimize PT100 temperatures
-
@dc42 Any more clues on how to get a Temperature offset / calibration working for a PT100 daughter board with M308?
I didn’t find anything about the R component for PT100‘s in the documentation.Also any chance an option for an actual (and simple) temperature offset will be added in the future?
-
@benecito said in Improve / optimize PT100 temperatures:
with M308?
I didn’t find anything about the R component for PT100‘s in the documentation.https://duet3d.dozuki.com/Wiki/M308
See the documentation for M308
Additional parameters for MAX31865-based PT100 sensors Rnnn Series resistor value. Leave blank to use the default for your board. Fnn (where nn is 50 or 60) The local mains frequency. Readings will be timed to optimise rejection of interference at this frequency.
-
@phaedrux It's been a while and because of various reasons this has become more relevant again.
I tried again and it does not really work for me. So far I usedM308 S1 P"spi.cs1" Y"rtd-max31865" A"Nozzle"
While this gives me a reading of 35.1°
R400 gives me approximately the same R401 gives me 35.9° and R399 gives me 34.8°
So not really an option for fine tuning the temperatures.
Any other ideas?I would still love to see an option to make this easier (and also think it would make sense in a lot of cases).
For example an "O" parameter in M308 that simply accepts offsets in a way like "O0.6" or "O-1.1" for corrections of 0.6° and -1.1°. -
Does the PT100 daughter board even have a series resistor?
-
@benecito the PT100 daughter board has a 400 ohm reference resistor. So a value of R400 should give the same reading as without the R parameter. You can use fractional values such as R400.1. Very roughly, the reading in deg Kelvin will be proportional to the R value.
The maximum temperature deviation caused by the MAX31856 chip should be no more than 0.5C according to the datasheet. The reference resistor has 0.1% tolerance so the maximum error it introduces should be no more than about 0.25C. Hence the total error should be no more than 0.75C plus the error introduced by the tolerance of the PT100 and any relevant wiring resistance.
-
@dc42 thanks for the reply! If there is a 400 ohm resistor at least that part makes sense
I just tried using fractional values, but it did not work for me. Everything from R400 to R400.9 gives me the same reading with a change of about 0.7° if I go up to R401.
Did this change since RRF3.3?Overall I am still very happy with the precision I get within one printer! I'd just like to get all the printers I build to be as close to each other as possible. Right now they are within 1° of each other, but with chocolate that already makes quite a difference, so ideally I'd like to get them within 0.2° of each other and thus be able to ship them calibrated.
-
@benecito you are correct, the R value for PT100 sensors accepts integer values only. I will change this in RRF 3.5beta1, possibly even in RRF 3.4.2.
-
@benecito I've put candidate 3.4.2beta1 binaries including this fix at https://www.dropbox.com/sh/iic3wmqb1m0pj7r/AACh921ueffhGhhDR07z7M8Ea?dl=0. If you wish to use them, please upgrade to release 3.4.1 first, then when you have that working you can update to 3.4.2beta1.
-
@dc42 Good to know I'm not too stupid to use it, and thanks for sharing the beta!
I'll give it a try as soon as I find a little time!Sorry if I have to ask again, but wouldn't it be much easier vor everyone to be able to tune the temperatures by applying an actual offset instead of fiddling with ADC offsets or resistor values? Also I think this should not be very hard to implement.
-
@dc42 Just tried it and it works. Better than nothing but I'd still prefer the easier option!
-
@benecito said in Improve / optimize PT100 temperatures:
Sorry if I have to ask again, but wouldn't it be much easier vor everyone to be able to tune the temperatures by applying an actual offset instead of fiddling with ADC offsets or resistor values?
If there is an error in the reading, then the best way to correct it depends on the nature of the error. If the reference resistor isn't quite the right value, or the PT100 element isn't quite the right value, then adjusting the R parameter is the correct solution; because adding an offset would lead to reading errors at higher temperatures.