ToolBoard: 2 wires fan, heater fault and how to extrude ?
-
can you measure the resistance of R21 on the boards
another thing to try is switch to the second thermistor port
121.temp1
-
Okay thank you I will try that this evening.
-
okay so the R21 resistance is about 2K18 for both two board.
I saw on an other post that is the correct value. -
then i dont know where the difference comes from.
you could get some pt1000 ones. maybe they provide better results. -
@Veti said in ToolBoard: 2 wires fan, heater fault and how to extrude ?:
then i dont know where the difference comes from.
you could get some pt1000 ones. maybe they provide better results.but the thermistor isn't the problem, the two have the same resistance.
-
@Veti I there a way to have the value that read the borad directly ?
-
i dont know where it could come from.
i would play a bit with the H and L values.
what do you get when you set the same on both?
-
@Veti I have a difference too, but I played a bit with the H and L value and now it is okay, the hotend doesn't return fault.
-
@Pseud3mys said in ToolBoard: 2 wires fan, heater fault and how to extrude ?:
I check the 2 resistances values, both are at 4.34 Kohm (~11°C). but one thermistor return ~10°C degrees more than the real temperature (~21°C)..
And it is enough to raise a fault with the heater, so it is disturbing.M308 S1 P"121.temp0" Y"thermistor" T500000 B3600 C1.196220e-7 H-12 L0 M308 S2 P"122.temp0" Y"thermistor" T500000 B3600 C1.196220e-7 H-10 L-12
All the value are equal exept for H and L but they don't change anything significantly (I tried to put the same value).
So why one of the thermistor return a wrong value ?You don't need to set the H and L values in config.g because after you run the semi-automatic calibration procedure, the required H and L values are store in non-volatile memory.
You could try removing the H and L values and run the semi-automatic calibration again on the board that is showing the error. But if that doesn't work, use your manually-adjusted H value.
-
@dc42 I change the H value manually and now it work. Thank you for your help !
-
@dc42 I have same problem. After few hours of printing the extruder heater goes offline, temperature goes to 2000 °C (I'm using PT1000)
Updated to latest 3.2-beta4.1 and problem remains. It keeps printing so I guess it's still keeping 220 °C, but to CAN bus it reports offline status.
M122 on Toolboard says:Diagnostics for board 1: Duet TOOL1LC firmware version 3.2-beta4.1 (2020-12-03) Bootloader ID: not available Never used RAM 4120, free system stack 36 words HEAT 46 CanAsync 89 CanRecv 83 TMC 54 MAIN 216 AIN 64 Last reset 11:35:12 ago, cause: software Last software reset at 2020-11-22 15:02, reason: HardFault, available RAM 4168, slot 0 Software reset code 0x0060 ICSR 0x00000003 SP 0x200011c8 Task Stack: 20001018 20001304 0000000f 00000001 20000928 000172cd 00018250 01000000 00077dfa 00077ef4 20001218 200032d0 00000020 00017a61 20004660 00000000 20001300 00008895 00077dfa 20004670 00077ef4 42651000 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 Driver 0: position 11794334, 409.0 steps/mm, standstill, SG min/max 0/112, read errors 0, write errors 1, ifcnt 35, reads 15910, writes 16, timeouts 0, DMA errors 0 Moves scheduled 562468, completed 562468, in progress 0, hiccups 0 No step interrupt scheduled VIN: 24.2V MCU temperature: min 39.2C, current 41.7C, max 51.2C Ticks since heat task active 43, ADC conversions started 41546663, completed 41546662, timed out 0 Last sensors broadcast 0x00000002 found 1 47 ticks ago, loop time 0 CAN messages queued 501025, send timeouts 0, received 938189, lost 0, free buffers 36
Any thoughts? Can I contribute with more debug infos to solve this?
-
@flajzi, it looks as though the heat task stack may still be too small. Please try the tool board firmware at https://www.dropbox.com/sh/jv86vnaqjs1j4qd/AAA00AGN-pHCX4ok4L81ukv-a?dl=0. Post a fresh M122 B1 report when either the problem occurs again or the print finishes.
-
@dc42 Hi again. Problem with 2000 °C is back.
It's keeping same temperature, but it reports nonsense. This is M122 output:Diagnostics for board 1: Duet TOOL1LC firmware version 3.2-beta4.1+1 (2020-12-12) Bootloader ID: not available Never used RAM 3960, free system stack 36 words HEAT 86 CanAsync 89 CanRecv 83 TMC 54 MAIN 209 AIN 64 Last reset 04:39:13 ago, cause: software Last software reset at 2020-11-22 15:02, reason: HardFault, available RAM 4168, slot 0 Software reset code 0x0060 ICSR 0x00000003 SP 0x200011c8 Task Stack: 20001018 20001304 0000000f 00000001 20000928 000172cd 00018250 01000000 00077dfa 00077ef4 20001218 200032d0 00000020 00017a61 20004660 00000000 20001300 00008895 00077dfa 20004670 00077ef4 42651000 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 Driver 0: position 7738615, 409.0 steps/mm, standstill, SG min/max 0/112, read errors 0, write errors 1, ifcnt 42, reads 53596, writes 14, timeouts 0, DMA errors 0 Moves scheduled 219075, completed 219075, in progress 0, hiccups 0 No step interrupt scheduled VIN: 24.2V MCU temperature: min 38.4C, current 54.5C, max 57.9C Ticks since heat task active 151, ADC conversions started 16686702, completed 16686701, timed out 0 Last sensors broadcast 0x00000002 found 1 155 ticks ago, loop time 0 CAN messages queued 201492, send timeouts 0, received 370230, lost 0, free buffers 36
-
@dc42 And hello again... Problem with 2000 °C persists also in RC1 release...
I don't know if that could be an issue, but I'm using PT1000 sensor instead of thermistor...Diagnostics for board 1: Duet TOOL1LC firmware version 3.2-RC1 (2020-12-19) Bootloader ID: not available Never used RAM 3960, free system stack 38 words HEAT 86 CanAsync 89 CanRecv 83 TMC 54 MAIN 219 AIN 64 Last reset 03:17:23 ago, cause: power up Last software reset at 2020-11-22 15:02, reason: HardFault, available RAM 4168, slot 0 Software reset code 0x0060 ICSR 0x00000003 SP 0x200011c8 Task Freestk 4294967295 ok Stack: 20001018 20001304 0000000f 00000001 20000928 000172cd 00018250 01000000 00077dfa 00077ef4 20001218 200032d0 00000020 00017a61 20004660 00000000 20001300 00008895 00077dfa 20004670 00077ef4 42651000 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 a5a5a5a5 Driver 0: position 7455393, 409.0 steps/mm, standstill, SG min/max 0/112, read errors 0, write errors 0, ifcnt 16, reads 23696, writes 16, timeouts 0, DMA errors 0 Moves scheduled 218054, completed 218054, in progress 0, hiccups 0 No step interrupt scheduled VIN: 24.2V MCU temperature: min 36.8C, current 51.7C, max 62.4C Ticks since heat task active 198, ADC conversions started 11796908, completed 11796907, timed out 0 Last sensors broadcast 0x00000002 found 1 202 ticks ago, loop time 0 CAN messages queued 142406, send timeouts 0, received 324917, lost 0, free buffers 36
My setup is this:
M308 S1 P"1.temp0" Y"pt1000" A"Nozzle"; configure sensor 1 as PT1000 on pin 1.temp0 M950 H1 C"1.out0" T1 ; create nozzle heater output on 1.out0 and map it to sensor 1 M143 H1 S285 ; set temperature limit for heater 1
-
Please post your complete config.g file.
Does the 2000C error occur immediately after starting the system, or is the displayed temperature normal at first and changes to 2000C and heater offline later? If it occurs immediately, try putting a G4 P500 command right at the start of config.g.
-
I have had the same issue. But I found out that screw in terminal block was broken, so i just soldered thermistor wire directly to PCB and problem dissapeared. So check wires and terminals.
-
@dc42 It always happens after few hours of printing. It never happens at the beginning of printing.
-
@Meltex @dc42 My issue is propably different. When it fails, toolboard is always keeping correct temperature at nozzle. It normally continues printing, toolboard is only reporting wrong temperature to Duet motherboard.
When I reset toolboard (unplug/plug power to whole printer), after startup it shows last correct temperature, few degrees below last defined temperature (hotend is cooled few degrees during restart).
So, PT1000 sensor is all time (even when toolboard reports fault to Duet board and 2000 °C) reporting correct temperature to the toolboard. -
@dc42 Sorry, I forgot post my config file.
; General preferences G90 ; send absolute coordinates... M83 ; ...but relative extruder moves M550 P"duet" ; set printer name M669 K1 ; select CoreXY mode ; Network ;M552 P0.0.0.0 S1 ; enable network and acquire dynamic address via DHCP ;M586 P0 S1 ; enable HTTP ;M586 P1 S0 ; disable FTP ;M586 P2 S0 ; disable Telnet ; Drives - M569 P0.0 S1 ; physical drive 0.0 goes forwards M569 P0.1 S1 ; physical drive 0.1 goes forwards M569 P0.2 S0 ; physical drive 0.2 goes backwards M569 P0.3 S0 ; physical drive 0.3 goes backwards M569 P0.4 S0 ; physical drive 0.4 goes backwards M569 P1.0 S0 ; physical drive 1.0 goes backwards M584 X0.0 Y0.1 Z0.2:0.3:0.4 E1.0 ; set drive mapping M350 X32 Y32 Z64 E16 I1 ; configure microstepping with interpolation M92 X160.00 Y160.00 Z1600.00 E409.00 ; set steps per mm ; Calculated values from https://www.3dblog.cz/nastaveni/akcelerace-jerk-a-junction-deviation/ M201 X3200.00 Y3200.00 Z100.00 E2000.00 ; set accelerations (mm/s^2) M203 X12000.00 Y12000.00 Z300.00 E10000.00 ; set maximum speeds (mm/min) M204 P1400 ; Set printing acceleration M205 X12.00 Y12.00 Z12.00 E12.00 ; set maximum instantaneous speed changes (mm/min) M906 X1773 Y1773 Z1773 E940 I30 ; set motor currents (mA) and motor idle factor in per cent M84 S30 ; Set idle timeout ; Axis Limits M208 X-4:270 Y-4:270 Z0:270 ; set axis minima/maxima ; Endstops M574 X1 S1 P"!1.io1.in" ; configure sensorless endstop for low end on X M574 Y1 S1 P"!io1.in" ; configure sensorless endstop for low end on Y M574 Z1 S2 ; configure Z-probe endstop for low end on Z ; Z-Probe M950 S0 C"1.io0.out" ; create servo pin 0 for BLTouch M558 P9 C"^1.io0.in" H1.5 F240 T12000 ; set Z probe type to bltouch and the dive height + speeds G31 P1000 X36 Y-44 Z4.50 ; set Z probe trigger value, offset and trigger height M557 X32:270 Y0:226 P8 ; define mesh grid ; Heaters M308 S0 P"temp0" Y"thermistor" T90000 B4000 A"Bed" D600 ; configure sensor 0 as thermistor on pin temp0 M950 H0 C"out1" T0 ; create bed heater output on out0 and map it to sensor 0 M140 H0 ; map heated bed to heater 0 M143 H0 S130 ; set temperature limit for heater 0 to 130C M308 S1 P"1.temp0" Y"pt1000" A"Nozzle"; configure sensor 1 as PT1000 on pin 1.temp0 M950 H1 C"1.out0" T1 ; create nozzle heater output on 1.out0 and map it to sensor 1 ;M950 H1 C"out0" T1 ; (TUNING PURPOSES - M303 only) create nozzle heater output on out0 and map it to sensor 1 M143 H1 S285 ; set temperature limit for heater 1 ; Fan M950 F0 C"!1.out1+^out1.tach" Q1000 ; create fan 0 on pin 1.out1 and set its frequency (Blower) M106 P0 S0 H-1 B0.2 ; set fan 0 value. Thermostatic control is turned off M950 F1 C"1.out2+^out2.tach" ; create fan 1 on pin 1.out2 and set its frequency (Heatsink) M106 P1 S1 H1 T45 ; set fan 1 value. Thermostatic control is turned on ; Tools M563 P0 D0 H1 F0 S"Hemera" ; define tool 0 G10 P0 X0 Y0 Z0 ; set tool 0 axis offsets G10 P0 R0 S0 ; set initial tool 0 active and standby temperatures to 0C ; Custom settings are not defined ; Miscellaneous M575 P1 S1 B57600 ; enable support for PanelDue ; Resetting Z-probe M280 P0 S160 ; Alarm Release and Push-Pin UP M280 P0 S90 ; Send PWM channel 0 the S90 (angle) command ;M290 M290 R0 S-0.040 ; Baby Z-offset M572 D0 S0.025 ; Configure pressure advance ;M564 H0 ; allow movement of axes that have not been homed M501 ; Set the active parameters to those stored in sys/config-override.g on the SD card.