@mendelevium said in Thermisors showing wrong temperatures:
@dc42
So the thermistor is plugged directly into the board while the thermocouples into the daughterboard right?Yes, correct.
-
@dc42
Thank you the bed sensor is working and reading a reasonable temperature
the hot end temperatures are not well... One is reading 2000deg all the time. The other spikes at 2000deg for a split second then goes to a low unrealistic temperature. -
Are you using extension wires on the thermocouples?
-
@phaedrux
yes of course but i have over 2 m of wireing. may try to connect a thermocouple directly and see the reading... -
Thermocouples can be very particular about their wiring.
https://duet3d.dozuki.com/Wiki/Connecting_thermocouples#Section_Troubleshooting
-
@phaedrux
I have fitted two 0.01microfarad capacitors on each channel to earth as I had found on the forum however everything remained the same. Now I connected a new thermocouple to the board which is not touching anything and the problem remains. -
-
Please post a photo showing the wiring between the daughter board and the thermocouple that is not connected to anything.
-
If you connect a wire link between the daughter board thermocouple terminals instead of connecting a thermocouple, what reading do you get?
-
-
@dc42
when I connected a wire link the sensor the temperature reads 23 degrees and it is not spiking. -
@dc42
the problem here seems to be complex because i have built numerous 3d printers and never had this kind of issue. I have the duet fitted in a metal enclosure which has two large harting type metal clad plugs x 24 pin each. The connections from the board run to the harting around 30 cm long. The machine has two harnesses plugged into the control. The harnesses are around 1 meter long and there is around 1 to 1.5 meters of cables inside the machine. The issue is that they run along with limit switches and motor controls. So my solution is to substitute the cables of the thermocouples and with screened cables unable to pick up interference. Plus possibly also change the type of thermocouples I am using... Its a massive job to rewire and I do not want to do this in vain... What do you suggest? -
@mendelevium said in Thermisors showing wrong temperatures:
@dc42
the problem here seems to be complex because i have built numerous 3d printers and never had this kind of issue. I have the duet fitted in a metal enclosure which has two large harting type metal clad plugs x 24 pin each. The connections from the board run to the harting around 30 cm long. The machine has two harnesses plugged into the control. The harnesses are around 1 meter long and there is around 1 to 1.5 meters of cables inside the machine. The issue is that they run along with limit switches and motor controls. So my solution is to substitute the cables of the thermocouples and with screened cables unable to pick up interference. Plus possibly also change the type of thermocouples I am using... Its a massive job to rewire and I do not want to do this in vain... What do you suggest?What sort of wire are you using to extend the thermocouples?
-
@mendelevium said in Thermisors showing wrong temperatures:
So my solution is to substitute the cables of the thermocouples and with screened cables unable to pick up interference.
Thermocouple wires carry very low voltages at low impedance. This makes them susceptible to inductive interference pickup. Screened cables are effective at shielding the cable from capacitive interference pickup, but much less effective at shielding from inductive pickup unless they are shielded twisted pair.
Thermocouple wires MUST always be twisted pair, and should be kept away from stepper motor cables. If it's impossible to keep them away from stepper motor cables, use twisted pairs for the stepper motor cables too, one pair per phase. To get accurate temperature readings, the thermocouple extension cables must be made for the same type of thermocouple wire as the thermocouple. See https://www.eurotherm.com/en/support-articles-en/traps-and-colour-confusion-in-thermocouple-wiring/.
-
@dougal1957
ordinary wires.... Never encountered this problem even when running S**T electronics which I would expect far more issues.... -
@mendelevium See DC42's reply above I was thinking the same as he was but wasn't 100% sure.
-
@dc42
interesting... are shielded and twisted wires found on the shelf?
do they have to be of a particular guage? -
@mendelevium said in Thermisors showing wrong temperatures:
@dc42
interesting... are shielded and twisted wires found on the shelf?
do they have to be of a particular guage?If the cable isn't the proper thermocouple extension wire, then that moves the cold junction from the daughter board to where the join is, resulting in inaccurate temperature readings. The reading will be off by approximately the difference between the temperature at the join and the temperature at the daughter board. See the link in my previous post.
Twisted pair is more important than shielding, however if the thermocouple wires run close to stepper motors wires then you may need both. The simplest option would be to use shielded twisted pair for the stepper motor cable (with the shield connected to Duet ground) and twisted thermocouple wire for the thermocouple.
Shielded twisted pair cable is readily available, for example Cat6a Ethernet cable.
-
@dc42
I have tried something. I have taken the loose thermistor and twisted the cable. Its just about 1 meter long. Connected this to the daughter board and pulled the wire straight up in a way that it is neither near nor in contact with anything. There is no change.. mostly spiking up to 2000...and touching it seems to invoke no temperature rise....
what do you suggest? Changing the thermistor to another brand.... i have around 100 pieces and its all the same behaviour
. -
@mendelevium, if it's a loose thermocouple then it should have been supplied with either a twisted cable, or a shielded twisted cable.
Is it definitely a thermocouple?
-
@dc42 this is tge thermocouple and the cable. Worked great with ramps/arduino.
-
@dc42 Normally i sleeve them in a brass micro canister 3mm dia with thermal paste
-
@mendelevium, I don't think that's a thermocouple. I think it is a thermistor. Can you measure its resistance using a multimeter?